Refresh
I am working with the last version to be sure (downloaded today : 3.0.20211123) on windows. There are refresh problems. When i change mA, CTDIvol doesn't change. If i want it to change, i change TCM strength from 0 to 0.
I am working with the last version to be sure (downloaded today : 3.0.20211123) on windows.
There are refresh problems.
When i change mA, CTDIvol doesn't change. If i want it to change, i change TCM strength from 0 to 0.25 (CTDIvol changes) and then go back to TCM 0.0. And the CTDIvol is the real good one.
Other problem, more important : changing the CT model, i can obtain 2 different nCTDIw !!! Once 6.8 and after changing a few times : 9 mGy. See screenshot.
Already refresh problems a guess ?
NCI Dose Team response
Thanks for reporting these errors in NCICT. We suspect the problem has something to do with the decimal separator (comma or dot) on your Windows since we see nCTDIvol values on your screenshots are 9, and 6,8 which are supposed to be shown as 9. and 6.8. However, we have never received the error reports like this... so we're not sure if this is general problems across all European users or only on your system. How about other programs NCINM or NCIRF? Are these working fine? Also can you try to install NCICT on your other Windows PC and see if you experience the same problem?
Community reply
Thank you for your feedback the NCI Dose Team,
I have change coma by dot, but still nCTDIw and refresh problem !
Gonna try on an other windows computer.
Didier